ASCII output format with data recording

by Sven Nylund last modified Mar 27, 2006 10:25 AM
Up to System integration and telemetry

ASCII output format with data recording

Posted by Sven Nylund at March 27. 2006
Quote
In the System Integrator Guide, rev. F, it states in the Continental ASCII Output description: "Observe that there is no storage of data to the internal recorder when data are output in ASCII format."
Does this apply to the Aquadopp and Aquadopp Profiler as well?


Since this revision of the manual this has been changed. In the latest firmware versions for both the Aquadopp, Aquadopp Profiler and Continental, it is possible to output data in ASCII format and store the data to the internal recorder at the same time.

Use the command RA for this. If no filename has been set the recording will start with DEF as the filename. If you want to set the filename you can start and stop a measurement from the software using Start with Recorder.

Best regards,

Sven Nylund