Troubleshooting
'> a) Spend a few minutes checking all the dumb things - make sure that the serial port is connected to the PC (make sure it is the right port!
, that the power connectors are all properly seated, that the underwater connector is properly mated, etc. b) Check the troubleshooting section of the manual. Among other things, it explains how to test the computer serial port.
c) The most common culprit is the power supply. Measure the voltage output but don't stop there - some times the power supply is not working properly and cannot supply the necessary current. In general, it is a great idea to measure the current with an amp-meter or with a lab supply. If the current is zero, the power does not reach the electronics and the failure could be in the cable or maybe the internal connector is not properly attached.
d) Be aware of the hardbreak/softbreak issue. Nortek changed from one type of "wake-up!" signal to another in 2001/2002 (exact time varies with the instrument) and you may have to toggle the break setting in the Communication / Serial port menu to change from one type of break to another. Use the terminal emulator window to check for the desired effect; the response from the instrument may be garbled if the baud rate is wrong but something will always show up on the screen if you send a break and the electronics is online.
'> , one I can't
'> . Same computer, software, comms cable, power supply.I've used the default BREAK that is initiated from the COMMUNICATION/Terminal Emulator dialog in Vector v1.25 software, with no response from the 'mute' Vector.
Are there any other steps I can take to restore comms?
Thanks,
-Chris
The first thing I normally check in a case like this is if the units draws any current. If you put an amp meter in series with the power supply line (or if you have a DC power supply that shows you the current) you can see if the unit is "dead" or not.
If it is "dead", it is time to open it up and take a look inside.
Best regards, Atle Lohrmann
Thank you for the reply. Directly upon applying 12VDC through the comms cable, with a multimeter (10A fused scale, dial set to 10A) wired in series to my Vector, it reads 0.03A (max precision.) This is about what I see from my power supply (varies from 0.02-0.04).
I don't know about 'death', but is rigor mortis setting in? Kidding aside, I still cannot communicate with the ADV. What did you mean by 'opening it up.' I have already opened the battery/comms side of the case to verify no water leakage, wire-loom intact, etc. Are you suggesting opening the probe-side of the housing?
Thanks,
-Chris
These are typical current drains
Command mode 8-12mA
Power down 1-2mA
Wait mode 2-4mA
Measurement mode 50-150mA
Your reading is something in between. Which power level was the Vector deployed with last time?
Please open the teminal emulator and check if it comes any characters at all. Turn det power on and off and see it anything displays then.
Are you running RS232 or RS422?
Best Regards
Jonas Røstad
The last time I attempted to deploy the Vector (when the instrument ceased to communicate) it was running from a 11.7V lithium battery.
I am using RS232 to communicate with the instrument. I have cycled power (numerous times) while using the terminal emulator. I get absolutely no character response in the window (despite seeing that the unit is drawing current.)
Again, when Vector v1.25 tries to autoconnect (cycling through all baud configs), it is unable to detect the instrument (baud.)
Thank you for the help,
-Chris
Ooops! By 'power level' did you mean from the dialog on the Advanced settings tab in Deployment Planning? If so, I configured it for "HIGH".
Thanks,
-Chris
This is starting to sound like a real hardware failure.
'>My last suggestion is that you try communicationg with it via RS422 (if you have the harness and cable available). If not - please contact inquiry@nortek.no for an RMA-number for returning the instrument for repair.
Best Regards
Jonas Røstad
Thanks,
-Jon
When receiving garbled responses you are right when suspecting wrong baudrate.
The software has a baudrate scanner if you choose Stop data collection in the instrument softwares main menu.
The voltage levels of the RS232 signal should be +/- 6,6V.
Best Regards
Jonas Røstad
Buen día señores Nortek,
Tengo el mismo problema de Chris.
El adcp no reponde a los comandos de break desde el terminal emulator y al autoconnect.
Probé configurando todos los baudrates y tampoco responde, aveces responde con una ÿ.
Abrí el equipo, mido tensiones y parecen estar correctas. Power supply = 18V, como indica el manual.
Tengo colocado un analizador de señales en el puerto RS232 de comunicaciones y veo que los datos son enviados hacia el adcp pero no se recibe nada. Aveces pone energía en el pin TX durante 5 minutos y luego se apaga. Es como que se enciende, espera a ser interrogado y como no recibe la cadena de datos correspondiente se vuelve al estado sleep y no vuelve al estado wake up.
¿Es posible que haya otra forma de enviarle comandos para que haga un wake up, por ejemplo usando el hyperterminal de windows?
Gracias, saludos.
Andrés Dorta
YPF / Mar del Plata / Argentina
-------------------
Nortek good day gentlemen,
I have the same problem for Chris.
The ADCP Repond not to break the commands from the terminal emulator and the AutoConnect.
I tried setting all baudrate and neither responds, sometimes responds with a ÿ.
I opened the computer, measure voltages and appear to be correct. = 18V power supply, as the manual.
I have placed an analyzer signals in the port RS232 communications and I see that the data are sent to the ADCP but did not receive anything. Sometimes putting energy into the pin TX for 5 minutes and then shuts down. It's like going on, waiting to be interrogated and have not received a string of data corresponding to the state becomes sleep and not wake up again to the state.
Is it possible that there may be another way to send commands to make a wake up, for example using the Windows HyperTerminal?
Thank you, greetings.
Andres Dorte
YPF / Mar del Plata / Argentina