Pings per sample

Up to HR-Profilers

Pings per sample

Posted by Eric Siegel at June 24. 2008

Can you please describe how to use the HR Profiler software to determine how many pings will be sent per sample?  Is the answer pings/beam or total pings?  What does the un-titled box just to the right of the "measurement load" describe?

Many thanks,
Eric

Current state: Being created

Re: Pings per sample

Posted by Atle Lohrmann at June 27. 2008

Hi, there -- this is a bug.  The number that you see in the greyed out box should be the measurement load that is actually used during the data collection but it is not properly scaled with the number of beams that are used.  It will get fixed.   -- Atle Lohrmann

Current state: Being created

Re: Pings per sample

Posted by Eric Siegel at July 02. 2008

More information about Measurement Load and Pings per Sample:

The measurement load works much the same way as for the standard Aquadopp Profiler.  In the HR mode, the maximum internal ping rate is 120 pings (for 2 MHz) distributed over the number of configured beams, hence 40 pings for the regular 3 beams. The grey field to the right of the measurement load is supposed to show the actual measurement load, but there is a slight problem in version 1.03 of the software. The problem is that it does not show the correct value for the one and two beam configuration, which will be fixed in the next release.  If using the 1 or 2 beam mode, you have to divide by 3 and multiply by the configured number of beams to get the correct value:

MeasLoad = oldMeasLoad*nBeams/3.

The maximum internal ping rate for the 1 MHz systems is a total of 59 pings which rounds down to 19 pings for the 3 beam configuration.

If you use the extended velocity range the software displays half the measurement load since half of the pings are used for the short lag.

 