Measurement load and data precision

Up to Deployment

Measurement load and data precision

Posted by admin at March 17. 2005
Quote
I want to measure puv data at 1Hz continuously for at least one week (before running out of battery/memory). What is the effect of decreasing the measurement load on the data?

The effect is two-fold. By increasing the measurement load, the precision in the velocity measurements increase, but alas at the expense of battery consumption. By decreasing the measurement load it is the other way around.

The exact relationship is shown in the Advanced tab in the Deployment planning menu. Just try and vary the measurement load, click on Apply and watch the change in the Battery utilization (for a specified assumed measurement duration) and the velocity precision. As you will see this is a trade-off between need for precision and need for duration.

A 100% measurement load means that the instrument pings as fast as it is able to. Consequently, a measurement load of 50% means that the instrument pings at half that rate.

From signal theory we know that the more pings there are within an averaging period, the better the estimate of the true value we're measuring will get.

A high ping rate will thus reduce the standard deviation, (which is called Precision in the Deployment planning menu).

To illustrate what we are talking about, consider the below list which provides the maximum internal sampling rate of some of our instruments:
 
Aquadopp profiler
2 MHz:23 Hz
1 MHz: 7 Hz
0.6 MHz: 4Hz

AWAC
1 MHz: 7 Hz
0.6 MHz: 4Hz

Continental
470 kHz: 5 Hz
190 kHz: 2 Hz

Gustav

Measurement load and data precision

Posted by admin at March 18. 2005
As a follow questions to Gustav's post on measurement load and battery utilization, what type of battery does the battery utilization function assume you are using, i.e. small internal alkaline - internal lithium - external large alkaline pack? Also, does the battery utilization function assume the battery is new at the time of programming? and if so are there any ways of estimating battery life through the battery utilization function if the battery is not new, especially the lithium type since they generally do not show a drop in voltage after being used?

Measurement load and data precision

Posted by admin at March 21. 2005
Dear ccimaglia,

The assumed type of battery is the one that is supplied with the instrument. Its total watthour [Wh] capacity is listed in the Deployment planning menu of the setup program that comes with your Nortek instrument.

The program assumes that your battery is fully charged and that it behaves like a new battery.

The simplest way of estimating what's left in the battery is to examine your previous setup(s) and the deployment time(s) elapsed. What's left should then be 100% minus the percentage already spent. Other ways of doing this will easily get complicated without yielding much more accuracy than you get by this simple method. <img src=" />:)'>

- gustav

Re: Measurement load and data precision

Posted by Sean Kilgallin at March 25. 2011

Hi,

Is internal sampling frequency per beam or for the entire instrument, so for a 2mHz at 100% load is each beam pinging at 23Hz or 23/3 (~8 ) Hz?

Thanks,

Sean

Re: Measurement load and data precision

Posted by Jonas Røstad at March 28. 2011
Hi Sean,
 
The number of pings are for the entire instrument for the AquaPro. 100% load is 23/3 (~8 ) Hz.

Best regards
Jonas Røstad

Re: Measurement load and data precision

Posted by Jonas Røstad at May 04. 2012

Hi again Sean,

 

One year later I reviewed my answer to you and found that it was completely wrong. 23Hz is per beam.

 

Best regards

Jonas Røstad

Re: Measurement load and data precision

Posted by Jonas Røstad at May 04. 2012

Since we have added an AWAC and an Aquadopp Profiler an updated table is needed:

 

Aquadopp profiler 
2 MHz:23 Hz
1 MHz: 7 Hz
0.6 MHz: 4Hz
0.4MHz: 3Hz

AWAC 
1 MHz: 7 Hz
0.6 MHz: 4Hz
0.4 MHz: 2Hz

Best regards

Jonas Røstad